Prof. Stoll,
I upload the original files and my M-file with the simulations.
Thank you very much!
I apologize for the delay.
Magnettech MS5000 transient measurements - wrong x-axis
- User
Re: Magnettech MS5000 transient measurements - wrong x-axis
- CoTMOP_77K_sim.m
- (3.22 KiB) Downloaded 1106 times
- User
Sorry, here are the spectra in a zipped file.
- CoTMOP_R27.zip
- (1.07 MiB) Downloaded 997 times
The xml file you sent contains several different data vectors:
- BField (2070 points)
- MW_Absorption (1550 points)
- MW_AbsorptionSinus (41473 points)
- MW_AbsorptionCosinus (41473 points)
Regarding the B field, the spectrometer stores the magnetic field axis raw data as a series of values sampled at some constant time interval (10 milliseconds or so). If the field gets high, the sweep is not perfectly linear, and this is reflected in the data. You can always generate a linear B axis via
eprload
loads MW_Absorption and interpolates the BField accordingly to give the field axis. You can also access the other two (which are the raw data, sampled every millisecond) viaCode: Select all
[B,spc,p] = eprload('data.xml');
spc1 = p.Curves.MW_AbsorptionSinus.data;
plot(spc1)
[B_,spc_,p] = eprload('data.xml');
B = linspace(B_(1),B(end),numel(B));
spc = interp(B_,spc_,B);
- User
I will from now on check the linearity always.
Nicolas
Nicolas
I tested the new version. The loading of the transient spectra works fine now.
Thanks for that!
However, now ES cannot load 2D measurements, such as saturation measurements, anymore. The error is
I think the problem is the detection of the measurement type which differentiates only between kinetic and single at the moment. If there is another (unknown) type it throws the above mentioned error. Other types are: powerSweep, modulationSweep, temperature. The list is unfortunately not comprehensive since we do not have all extensions for the spectrometer (additional upgrades are a goniometer and an X-ray source). For a robust implementation it might be good to check only for the type "transient", which leads to the time x-axis.
Thanks for that!
There are no real 2D measurements possible with the MS5000. In addition to B0 another parameter can be swept, such as power, modulation amplitude, temperature. Each spectrum at a certain power/modulation amplitude/temperature/... is still saved in a single xml file with a different measurement type identifier (see also attached spectra).Error using eprload
Cannot construct abscissa for this measurement type
<Recipe Name="LOGSTest_008" Author="" Type="modulationSweep" MeasurementCount="1" CreationDate="2018-08-29T09:37:04.2687361Z">
<Recipe Name="LOGSTest_009" Author="" Type="powerSweep" MeasurementCount="1" CreationDate="2018-08-29T09:48:26.4967361Z">
<Recipe Name="MINST_090_SK1118" Author="" Type="temperature" MeasurementCount="1" CreationDate="2018-05-25T09:26:34.6064526Z">
- ES_Magnettech_2.zip
- (1.42 MiB) Downloaded 980 times
- ES_Magnettech_1.zip
- (2.76 MiB) Downloaded 1182 times
Thanks for this additional information and data. We will look into this.
Ideally, we just implement an official file format specification from Magnettech. Unfortunately, they are not able to provide us with one. Therefore, we need to improve
eprload
incrementally (and tediously) using example data.-
This is fixed now in 5.2.23.
It works like a charm now.
Thanks a lot!
Thanks a lot!