Magnettech MS5000 transient measurements - wrong x-axis

A place to report and discuss potential bugs
niconeuman
User
Posts: 11
Joined: Mon Oct 22, 2018 1:16 am

Re: Magnettech MS5000 transient measurements - wrong x-axis

Post by niconeuman »

Prof. Stoll,
I upload the original files and my M-file with the simulations.
Thank you very much!
I apologize for the delay.
Attachments
CoTMOP_77K_sim.m
(3.22 KiB) Downloaded 664 times
niconeuman
User
Posts: 11
Joined: Mon Oct 22, 2018 1:16 am

Re: Magnettech MS5000 transient measurements - wrong x-axis

Post by niconeuman »

Sorry, here are the spectra in a zipped file.
Attachments
CoTMOP_R27.zip
(1.07 MiB) Downloaded 576 times
Stefan Stoll
EasySpin Creator
Posts: 1108
Joined: Mon Jul 21, 2014 10:11 pm
Location: University of Washington

Re: Magnettech MS5000 transient measurements - wrong x-axis

Post by Stefan Stoll »

The xml file you sent contains several different data vectors:
- BField (2070 points)
- MW_Absorption (1550 points)
- MW_AbsorptionSinus (41473 points)
- MW_AbsorptionCosinus (41473 points)

eprload loads MW_Absorption and interpolates the BField accordingly to give the field axis. You can also access the other two (which are the raw data, sampled every millisecond) via

Code: Select all

[B,spc,p] = eprload('data.xml');
spc1 = p.Curves.MW_AbsorptionSinus.data;
plot(spc1)
Regarding the B field, the spectrometer stores the magnetic field axis raw data as a series of values sampled at some constant time interval (10 milliseconds or so). If the field gets high, the sweep is not perfectly linear, and this is reflected in the data. You can always generate a linear B axis via

Code: Select all

[B_,spc_,p] = eprload('data.xml');
B = linspace(B_(1),B(end),numel(B));
spc = interp(B_,spc_,B);
niconeuman
User
Posts: 11
Joined: Mon Oct 22, 2018 1:16 am

Re: Magnettech MS5000 transient measurements - wrong x-axis

Post by niconeuman »

Thank you very much!
I will from now on check the linearity always.
Best regards,
Nicolas
Silvio
User
Posts: 10
Joined: Mon Oct 16, 2017 8:37 am

Re: Magnettech MS5000 transient measurements - wrong x-axis

Post by Silvio »

I tested the new version. The loading of the transient spectra works fine now.

Thanks for that!

However, now ES cannot load 2D measurements, such as saturation measurements, anymore. The error is
Error using eprload
Cannot construct abscissa for this measurement type
There are no real 2D measurements possible with the MS5000. In addition to B0 another parameter can be swept, such as power, modulation amplitude, temperature. Each spectrum at a certain power/modulation amplitude/temperature/... is still saved in a single xml file with a different measurement type identifier (see also attached spectra).

Code: Select all

<Recipe Name="LOGSTest_008" Author="" Type="modulationSweep" MeasurementCount="1" CreationDate="2018-08-29T09:37:04.2687361Z">

Code: Select all

<Recipe Name="LOGSTest_009" Author="" Type="powerSweep" MeasurementCount="1" CreationDate="2018-08-29T09:48:26.4967361Z">

Code: Select all

<Recipe Name="MINST_090_SK1118" Author="" Type="temperature" MeasurementCount="1" CreationDate="2018-05-25T09:26:34.6064526Z">
I think the problem is the detection of the measurement type which differentiates only between kinetic and single at the moment. If there is another (unknown) type it throws the above mentioned error. Other types are: powerSweep, modulationSweep, temperature. The list is unfortunately not comprehensive since we do not have all extensions for the spectrometer (additional upgrades are a goniometer and an X-ray source). For a robust implementation it might be good to check only for the type "transient", which leads to the time x-axis.

Thank you very much!
Attachments
ES_Magnettech_2.zip
(1.42 MiB) Downloaded 599 times
ES_Magnettech_1.zip
(2.76 MiB) Downloaded 583 times
Stefan Stoll
EasySpin Creator
Posts: 1108
Joined: Mon Jul 21, 2014 10:11 pm
Location: University of Washington

Re: Magnettech MS5000 transient measurements - wrong x-axis

Post by Stefan Stoll »

Thanks for this additional information and data. We will look into this.

Ideally, we just implement an official file format specification from Magnettech. Unfortunately, they are not able to provide us with one. Therefore, we need to improve eprload incrementally (and tediously) using example data.
Stefan Stoll
EasySpin Creator
Posts: 1108
Joined: Mon Jul 21, 2014 10:11 pm
Location: University of Washington

Re: Magnettech MS5000 transient measurements - wrong x-axis

Post by Stefan Stoll »

This is fixed now in 5.2.23.
Silvio
User
Posts: 10
Joined: Mon Oct 16, 2017 8:37 am

Re: Magnettech MS5000 transient measurements - wrong x-axis

Post by Silvio »

It works like a charm now.

Thanks a lot!