Amp in resfield
Amp in resfield
Hi:
I am trying to use Amp produced by resfields function. May I know whether Amp is gamma times the square of matrix element or just the square of matrix element itself? For example, in the paper(http://e-collection.library.ethz.ch/ese ... 411-02.pdf), is Amp calculated in the same way as Equation (2.50)?
Thank you very much in advance for your help!
Re: Amp in resfield
resfields
includes in the amplitudes (a) the square of the transition magnetic dipole moment, (b) the field-to-frequency conversion factor (which you call gamma), and (c) if a temperature is given, the population difference according to the Boltzmann distribution.Re: Amp in resfield
Sorry for either hijacking or necrobumping this thread, but I would have chosen the same subject title for my question.
As far as I understand from your answer, @Stefan Stoll, the weighting factor of a given orientation is not taken into account in the amplitude output by resfields, is it? That means that if I would like to reconstruct the intensities from a spectral simulation, I also have to use the weights output by sphgrid. Or do I miss something here?
Re: Amp in resfield
That is absolutely correct.
Re: Amp in resfield
I am taking my turn to resurrect this question, as the thread is the most relevant: I would like to know how we can strip the
I would like to compare the transition probabilities calculated from ES with the ones analytically derived in different textbooks and papers for different exchange-coupled systems. For the arithmetic result to be directly comparable, should I take special care of the energy and field units?
There is also an undocumented
Thanks!
resfield
calculation down only to the matrix element |<n|Sx|m>|^2, ignoring all other factors.I would like to compare the transition probabilities calculated from ES with the ones analytically derived in different textbooks and papers for different exchange-coupled systems. For the arithmetic result to be directly comparable, should I take special care of the energy and field units?
Grad
field, which (when we set Opt.Verbosity to 2) is commented as gradient. Would that be the orientational gradient of a state energy?Thanks!
Re: Amp in resfield
Here is a simple demo script on how to reproduce the transition rates calculated by
resfields
. Obvioulsy, if you tilt the magnetic field relative to the g-tensor frame, then things get a bit more tedious due to the Euler angles needed to handle the tilted orientations, but the physics is exactly the same.Code: Select all
clc, clear
Sys.S = 1/2;
Sys.g = [2 2.1 2.2];
Exp.Range = [300 400];
Exp.CrystalOrientation = [0 0 0];
Opt.Freq2Field = false;
[B,TransRate0] = resfields(Sys,Exp,Opt)
H = sham(Sys,[0 0 1]*B);
[V,E] = eig(H);
[Sx,Sy,Sz] = sop(Sys.S,'x','y','z');
u = V(:,1);
v = V(:,2);
c = bmagn*Sys.g(1)/planck/1e9;
TransRate1 = c^2 * (abs(u'*Sx*v)^2 + abs(u'*Sy*v)^2)/2
Re: Amp in resfield
Thanks Stefan!
I had been reading the
Indeed, when I used this sample script to make the calculation for my system,
sop
documentation and couldn't figure out how to treat exchange-coupled systems. E.g., something like:
Sys.S = [5/2 5/2 5/2];
Gx = 0; Gy = 0; Gz = 0.2;
GG = [0 Gz -Gy; -Gz 0 Gx; Gy -Gx 0];
gz = 2;
gxy = 2;
Ja = -22.6;
Jb = -19.3;
Sys1.g = [gxy gz; gxy gz; gxy gz];
Sys1.ee = (100*clight/1e6)*[-2*Ja*eye(3) - 2*GG; -2*Ja*eye(3) + 2*GG; -2*Jb*eye(3) - 2*GG];
sop
threw an error:
Is there a way to get around this?Error using sop
sop: Could not determine what 'x' is for the given spin system.
Error in sop
Re: Amp in resfield
For calculating the transition moments and rates, you need the sum over spins:
[S1x,S1y,S1z] = sop(Sys.S,'x1','y1','z1');
[S2x,S2y,S2z] = sop(Sys.S,'x2','y2','z2');
[S3x,S3y,S3z] = sop(Sys.S,'x3','y3','z3');
Sx = S1x+S2x+S3x;
Sy = S1y+S2y+S3y;
Re: Amp in resfield
Thanks for this Stefan,
I now made the calculations as you suggested, assuming:
For the previously defined system, and for those experimental conditions, the results are:
<1|Sx|2> = <1|Sy|2> = 0.231365072811767
and since TransRate1 = c^2 * (<1|Sx|2> + <1|Sy|2>)/2 with c = bmagn*Sys.g(1)/planck/1e9 = 27.99249, I get Trans1Rate1 = 181.2929.
However the
UPDATE: Subsequently, I calculated the Boltzmann populations of states 1 and 2 and multiplied with their difference:
which yields 20.206058, about double the result from
Exp.Temperature = 4.2; Exp.mwFreq=9.801; Exp.CenterSweep=[350 700];
However the
resfields
result is TransRate0 = 10.1332 (with a very minor additional transition corresponding to transition 3-4). By removing the temperature from Exp
, resfields
calculates all possible resonances, and then indeed the 1-2 resonance is given with a TransRate0 = 181.2929 (and all others with non-realistic values for my experimental conditions). So I conclude that TransRate0 and TransRate1 coincide when we neglect thermal populations, right?UPDATE: Subsequently, I calculated the Boltzmann populations of states 1 and 2 and multiplied with their difference:
pop = exp(-Ener*1e6*planck/boltzm/Exp.Temperature);
Intensity1 = TransRate1 * (pop(1) - pop(2))
resfield
. I suppose I need to apply the 1/g Aasa-Vanngard correction. This gives practically the resfields
result if g = 2.-
Re: Amp in resfield
You got it!