ELDOR-detected NMR VS symmetry!!

General forum for EasySpin: questions, how to's, etc.
Post Reply
gabrielthomass
Newbie
Posts: 3
Joined: Mon Apr 29, 2019 1:39 am

ELDOR-detected NMR VS symmetry!!

Post by gabrielthomass »

Dear someone,

Recently I have been working on the simulation of EDNMR for an spin half and half powder system using the code provided by Anton.Savitsky. It turns out if I specified different symmetry in the simulation, the NMR intensity will be different. I checked the code, which use sphgrid function to generates triangular grids of orientations over parts of the unit sphere and symmetry specifies the symmetry for which the grid is to be generated,different symmetry will have different unique surface, which means in the simulation, they use different part of the unit sphere. But I am confused how the NMR intensity can be influenced by the choices of the symmetry (some symmetries will give the same intensity, but some will not).

I would be really appreciated if someone can help.https://redtube.onl/ https://beeg.onl/ https://spanktube.vip/spankbang/
Stefan Stoll
EasySpin Creator
Posts: 1100
Joined: Mon Jul 21, 2014 10:11 pm
Location: University of Washington

Re: ELDOR-detected NMR VS symmetry!!

Post by Stefan Stoll »

The higher the symmetry of the spin Hamiltonian, the smaller the surface of the unit sphere needed for obtaining a full powder spectrum. If in doubt, always use C1 or Ci symmetries - these are the lowest and work with any spin Hamiltonian.