Linewidths in mT and MHz
Posted: Tue Jul 13, 2021 4:10 am
by thanasis
Hello,
From the very basic relation between MW frequency and resonance magnetic field (see mt2mhz
/ mhz2mt
functions), one is proportional to the other.
I would therefore expect that line widths in MHz (for frequency-sweep experiments) and mT (for field-sweep experiments) would be interconvertible with the above functions.
Is that, however, rigorously correct? Is there some finer factor that has to be considered?
Thanks!
Re: Linewidths in mT and MHz
Posted: Wed Jul 14, 2021 10:45 am
by Stefan Stoll
It's not totally trivial. Several points:
- The more fundamental property is the frequency-domain line broadening. In EasySpin, it is assumed to be Gaussian. Whenever you use any of the strain broadenings, this gets used.
- EasySpin uses field-domain linewidths (
Sys.lwpp
and Sys.lw
) just for convolutional broadening. Therefore, every orientation gets the same broadening, even if the g tensor is very anisotropic. There is no physics behind these convolutional broadenings, i.e. no mechanism is assumed. It's purely phenomenological.
- In general, the field-domain line shape resulting from a Gaussian frequency-domain broadening is non-Gaussian and asymmetric, if the energy level dependence on the magnetic field is non-linear (which it is very often).
A good read regarding this topic is Pilbrow's book on Transition Metal EPR.