resfreqs - Using different orientations
Charles
Post by Charles »
I want to calculate the resonance frequencies in the x, the y and the z direction. I use e.g. the following code:
Sys.S = 1;
Sys.D = -30000;
Sys.g = 2;
Params.CrystalOrientation = [0 pi/2; pi/2 pi/2; 0 0];
Params.Field = 1000;
resfreqs_matrix(Sys,Params)
I expect that there are three columns as an output, of which two should have the same value because the system does not differ in the x and the y direction. But I only get this output:
ans =
   1.0e+04 *
    1.9948    1.9948
    7.1897    7.1897
    5.1948    5.1948
It looks like the z direction is missing. What am I doing wrong?
Re: resfreqs - Using different orientations
In Params.CrystalOrientation, EasySpin expects all three Euler angles for each orientation. So each row should have 3 values: phi, theta and chi. Since chi only affects the intensities, but not the resonance positions, you can use chi=0.
What is happening in your case is that EasySpin sees 3 rows (instead of 3 columns) and transposes the matrix to get the 3 columns for the 3 angles. As a consequence, you get something different from what you expected.
If you use
Params.CrystalOrientation = [0 pi/2 0; pi/2 pi/2 0; 0 0 0];
then you should get what you expect.